Introduction and Background
Dwell volume, Dead volume and Void volume play a vital role in chromatographic method development. Due to a lack of clarity analytical chemists do not use these terms effectively while developing the analytical method. That is why I decided to share my skill-based knowledge on Dwell volume, Dead volume, Void volume, Chromatogram, Peak, Unretained peak, Retention time and Relative retention time. Having read this post all your doubts will be cleared and you will be able to answer questions like:
The dwell volume is the volume of the mobile phase of the gradient HPLC system between the mixing chamber and column inlet. It is also called Gradient delay volume. It is denoted by VD. It is the primary function of the pump.
The following mobile phase (Table-1), chromatographic condition (Table-2), formula (equation-1 and equation-2) and gradient curve (figure -1) are used to calculate the Dwell volume.
Mobile phase gradient
Time (minute) | Mobile phase A – water (V/V) | Mobile phase A – 0.1% Acetonitrile (V/V) |
0 | 100 | 0 |
20 | 0 | 100 |
30 | 0 | 100 |
Chromatographic condition
Column | (1m x 0.12mm), PEEK capillary tubing can be used |
Flow rate | 2ml/minute |
Wavelength | 265nm |
Dwell volume (D) can be calculated by following formula
Where: D is the Dwell volume, F is flow rate and tD is the Dwell time
tD can be calculated by following formula:
Where:
Once the gradient is completed the chromatogram looks like as in figure-1. The initial flat baseline represents 0% acetone (100%A) and the final baseline is 0.1% acetone (100% B). The sloping line is the transition between the two. The following approaches are used to calculate the Dwell time (tD).
The Dead volume is the volume of the mobile phase of the HPLC system between point injection to point of detection. The volume of the void space of the column/column is not considered in the dead volume.
The Dead volume can be measured by replacing the column with a zero dead volume connector or union
Procedure
Where : t = time of injection and to = Peak elution time
In the HPLC column, some of the space between the stationary phase is not occupied by the stationary phase that space is called void space. The same void space can be occupied by solvent or mobile phase.
The volume of the solvent/mobile phase in a liquid chromatography column is called the Void volume. In another way, we can say that the void volume is the volume of the column not occupied by the support particles. It includes volume between the particles as well as volumes of the pores of the particle. It is denoted by V.
Void volume is calculated by the following formula:
Void volume Calculation (equation -4)
t0 = Void volume/Flow rate of the mobile phase
Where:
Case study: Suppose column dimension is (150 x 2.1)mm and f is 0.70. Then void volume will be:
V = 3.14159265359 x (2.1)2 x 150 x (0.70/4) = 0.37ml
V can also be calculated using following formula :
V = to x Flow rate of the mobile phase
Where to is the retention time (RT of the unretained peak
Void volume | Dead volume |
The volume of the solvent/mobile phase in a liquid chromatography column is called Void volume. In another way, we can say that the void volume is the volume of the column not occupied by the support particles. It includes volume between the particles as well as volumes of the pores of the particle. It is denoted by V | The Dead volume is the volume of the HPLC system between the point of injection and appointed detection without the column. |
Volume of the tubing is not considered | Volume of column is not considered |
The peak due to void volume is called void volume peak or unretained peak. it is denoted by to
to can be calculated using the following procedure -1 and procedure-2
t0 = Void volume/Flow rate of the mobile phase
Case study: Suppose column dimension is (150 x 2.1)mm and flow rate of mobile phase is 0.2ml/minute. Then to will be (0.30/2) = 1.85 minute
It is a two dimensional graphical representation of the concentration of the analyte/analytes. The x-axis represents time and the y – axis represents analyte concentration in terms of detector response.
When the analyte goes into the detector, the detector converts it into the signal and sends that signal to the data processor and the data process converts that signal into the peak.
The peak is a two-dimensional graphical representation of the concentration of the analyte in the chromatogram. The x-axis represents time and the y – axis represents analyte concentration in terms of detector response.The peak should be Gaussian or bell-shaped.
One chromatogram contains multiple peaks where one peak represents only one analyte or concentration of one analyte.
Case studies: If the sample mixture contains phenol and Benzoic then its chromatogram will contain two peaks; one for Benzoic acid and a second for Phenol
The time at which peak analyte elutes in the chromatogram as a peak is called Retention time. It is denoted by RT
It is the relative location of two peaks. Generally, the relative location of the impurity peak or any other peak is calculated against the main analyte peak. Hence we can say that Relative retention time is the ratio of the retention time of an impurity peak and the retention time of the main peak. It is denoted by RRT.
RRT = RT of impurity peak/RT of main peak
Chromatographic terms “Dead volume, Dwell volume, Void volume, chromatogram, peak, unretained peak, Retention time and Relative retention time are used in chromatographic analysis. development. Now I can hope that all your doubts have been cleared related to these terms. Please write your review or question in the comment section and I will answer the same on a priority basis.
You may also want to check out other articles on my blog, such as: