Dwell volume, Dead volume and Void volume

Table of Contents

Introduction and Background

Dwell volume, Dead volume and Void volume play a vital role in chromatographic method development. Due to a lack of clarity analytical chemists do not use these terms effectively while developing the analytical method. That is why I decided to share my skill-based knowledge on Dwell volume, Dead volume, Void volume, Chromatogram, Peak, Unretained peak, Retention time and Relative retention time. Having read this post all your doubts will be cleared and you will be able to answer questions like:

  • What is dead volume and how it is measured?
  • What is the dwell volume and how it is measured?
  • What is the void volume and how it is calculated?
  • What is the void volume peak or unretained peak and how it is calculated?
  • What is the chromatogram?
  • What is the peak?
  • What is the retention time?
  • What is the relative retention time?
  • What is the capacity factor and how it is calculated?

Dwell volume, Dead volume, Void volume, Chromatogram, Peak, Unretained peak, Retention time and relative retention time

Dwell volume and its functions

The dwell volume is the volume of the mobile phase of the gradient HPLC system between the mixing chamber and column inlet. It is also called Gradient delay volume. It is denoted by VD. It is the primary function of the pump.

  • It is responsible for delay for a gradient
  • It impacts on separation of different components in gradient mode
  • If two systems have different dwell volumes then there may be variation in retention time using the same chromatographic method/condition
  • It does not impact on separation in isocratic mode
  • Lower is the dead volume better is the instrument
  • It is not related to HPLC column and consequently Volume of the void space of the column is not considered in the Dwell volume.

Measurement of the Dwell volume

The following mobile phase (Table-1), chromatographic condition (Table-2), formula (equation-1 and equation-2) and gradient curve (figure -1) are used to calculate the Dwell volume.

Mobile phase gradient

Time (minute)Mobile phase A – water
(V/V)
Mobile phase A – 0.1% Acetonitrile
(V/V)
01000
200100
300100
Table-1

Chromatographic condition

Column (1m x 0.12mm), PEEK capillary tubing can be used
Flow rate2ml/minute
Wavelength265nm
Table-2

Dwell volume (D) can be calculated by following formula

D = tD x F (equation -1)

Where: D is the Dwell volume, F is flow rate and tD is the Dwell time

Dwell time (tD) calculation procedure

tD can be calculated by following formula:

tD = t50 -1/2 tG or tD = t50 -0.5 tG (equation -2)

Where:

  • G is 20 minutes (see gradient, table-1),
  • t0.5 is the time in minutes when absorbance has increased by 50% (see figure xx)

tD measurement procedure

Dwell volume, Dead volume and Void volume in HPLC
Image source: USP

Once the gradient is completed the chromatogram looks like as in figure-1. The initial flat baseline represents 0% acetone (100%A) and the final baseline is 0.1% acetone (100% B). The sloping line is the transition between the two. The following approaches are used to calculate the Dwell time (tD).

  • Approach-1: Draw a line through the gradient portion of the curve, extending below the baseline (dashed line figure – 1). Next, extend the initial baseline until it intersects the line. The intersection represents the Dwell time tD of the system. Now D can easily be calculated using equation -1. If tD is 0.85 then D will be; D = tD x F, D = 0.85 x 2 = 1.7
  • Approach 2: Measure the offset between the low and high baseline and find the midpoint. Extend a line from the midpoint horizontally until it intersects the rising baseline. From this intersection, drop a line vertically to the time axis. Subtract half of the gradient time (20 minutes/2 = 10 minutes in the present case). The remaining time is the Dwell time (tD). Now D can easily be calculated using equation -1 same as the above procedure (Approach-1)

Dead volume and its functions

The Dead volume is the volume of the mobile phase of the HPLC system between point injection to point of detection. The volume of the void space of the column/column is not considered in the dead volume.

  • It is applicable for both gradient and isocratic system.
  • It is not related to HPLC column and consequently volume of the void space of the column is not considered in the Dead volume
  • Lower is the dead volume better is the instrument

Measurement of Dead volume

The Dead volume can be measured by replacing the column with a zero dead volume connector or union

Procedure

  • Inject the the sample (at minimum concertation) and note down injection time
  • Note down the peak elution time

Dead volume = (t – to)x Flow rate (equation-3)

Where : t = time of injection and to = Peak elution time

Void volume

In the HPLC column, some of the space between the stationary phase is not occupied by the stationary phase that space is called void space. The same void space can be occupied by solvent or mobile phase.

The volume of the solvent/mobile phase in a liquid chromatography column is called the Void volume. In another way, we can say that the void volume is the volume of the column not occupied by the support particles. It includes volume between the particles as well as volumes of the pores of the particle. It is denoted by V.

Void volume in HPLC column

Void volume is calculated by the following formula:

Where:

  • V = Void volume
  • d = internal diameter of the column
  • l = length of the column
  • f = pore volume. It is 0.70 for fully porous packing and 0.50 for superficially porous packing

Case study: Suppose column dimension is (150 x 2.1)mm and f is 0.70. Then void volume will be:

V = 3.14159265359 x (2.1)2 x 150 x (0.70/4) = 0.37ml

V can also be calculated using following formula :

Where to is the retention time (RT of the unretained peak

Difference between Void volume and Dead volume

Void volumeDead volume
The volume of the solvent/mobile phase in a liquid chromatography column is called Void volume. In another way, we can say that the void volume is the volume of the column not occupied by the support particles. It includes volume between the particles as well as volumes of the pores of the particle. It is denoted by VThe Dead volume is the volume of the HPLC system between the point of injection and appointed detection without the column.
Volume of the tubing is not consideredVolume of column is not considered
Table-3

Void peak or unretained peak

The peak due to void volume is called void volume peak or unretained peak. it is denoted by to

Procedure for calculating Void volume peak or unretained peak

to can be calculated using the following procedure -1 and procedure-2

  • Procedure -2: to can be calculated using the following formulae:

Case study: Suppose column dimension is (150 x 2.1)mm and flow rate of mobile phase is 0.2ml/minute. Then to will be (0.30/2) = 1.85 minute

  • Procedure -2: to is also can also be calculated using unretained analytes like Uracil, KI… (in Reverse HPLC mode)
  • Very helpful in HPLC method development
  • Higher is the separation between unretained peak and peak of interest better is the method
  • It is helpful in calculating capacity factor

Chromatogram and peak

Chromatogram

It is a two dimensional graphical representation of the concentration of the analyte/analytes. The x-axis represents time and the y – axis represents analyte concentration in terms of detector response.

Peak

When the analyte goes into the detector, the detector converts it into the signal and sends that signal to the data processor and the data process converts that signal into the peak.

The peak is a two-dimensional graphical representation of the concentration of the analyte in the chromatogram. The x-axis represents time and the y – axis represents analyte concentration in terms of detector response.The peak should be Gaussian or bell-shaped.

Difference between the chromatogram and the peak

One chromatogram contains multiple peaks where one peak represents only one analyte or concentration of one analyte.

Case studies: If the sample mixture contains phenol and Benzoic then its chromatogram will contain two peaks; one for Benzoic acid and a second for Phenol

Retention time and Relative retention time

Retention time

The time at which peak analyte elutes in the chromatogram as a peak is called Retention time. It is denoted by RT

Relative retention time

It is the relative location of two peaks. Generally, the relative location of the impurity peak or any other peak is calculated against the main analyte peak. Hence we can say that Relative retention time is the ratio of the retention time of an impurity peak and the retention time of the main peak. It is denoted by RRT.

Chromatogram showing RT, RRT and Void volume peak
Chromatogram showing RT, RRT and Void volume peak

Conclusion

Chromatographic terms “Dead volume, Dwell volume, Void volume, chromatogram, peak, unretained peak, Retention time and Relative retention time are used in chromatographic analysis. development. Now I can hope that all your doubts have been cleared related to these terms. Please write your review or question in the comment section and I will answer the same on a priority basis.

You may also want to check out other articles on my blog, such as:

Abbreviations:

  • RT: Retention time
  • RRT: Relative retention time
  • Void volume: V
  • Dwell volume:
  • to : Unretained time

References:

  • IP
  • USP
  • BP

3 Responses to “Dwell volume, Dead volume and Void volume in HPLC

  • Shilpi Pandey
    1 year ago

    Lajawaab post. Very nicely and thouroughly explained. Regards

  • Shilpi Pandey
    1 year ago

    Lajawaab post. Very nicely and thoroughly explained. Regards

  • Manigandan g
    1 year ago

    Really wonderful post sir. Framed and explained well.

Leave a Reply

Your email address will not be published. Required fields are marked *

You cannot copy content of this page

×