How to calibrate FTIR?

Table of Contents

Calibration description

Following are the main sections for the calibration of FTIR

Objective

The purpose of this SOP is to detail the for the calibration of FTIR

Scope

This procedure applies to all FTIR instruments of xxxx (where xxx is the company name)

Scope

The procedure applies to all FTIR at QC /Analytical Laboratory of xxx (where xxx is the company name)

Responsibility

Research scientists/Chemists/Analysts responsible for the calibration of the FTIR

HOD

To ensure the implementation of this SOP in the department (QC or Analytical or as applicable)

FTIR calibration Frequency

Once in three months ± 5 days

Calibration procedure

  1. Operate the instrument as per operational SOP
  2. Make sure the instrument must be clean
  3. Use Polystyrene film reference standard with valid COA for calibration. The Thickness of the Polystyrene film should be 38 µm (micron) or 0.038mm
  4. Set the Polystyrene film in the sample chamber of the instrument and set the instrument parameters as mentioned below:
    • Resolution: 2 cm-1
    • Measurement mode: % transmittance ratio
    • Number of Scan: 1 to 20
    • Scan range: 4000 to 400 cm-1

Note: The scan range may be different for some of the instruments. Therefor scan range must be set as per Instrument requirement

Procedure for the verification of the Wavenumber

The wavenumber is verified using the calibrated Polystyrene film, which has transmittance minima (absorption maxima) at wavenumbers (in cm-1) shown below table – 1:

Table-1

Std Wavenumber(cm-1)Tolerance(cm-1)Acceptance criteria(cm-1)
3060±13059 to 3061
2849.5±12848.5 to 2850.5
1942.9±11941.9 to 1943.9
1601.2±11600.2 to 1602.2
1583±11582 to 1584
1154.5±11153.5 to 1155.5
1028.3±11027.3 to 1029.3
FTIR Verification of the Wavenumber

Procedure for control of Resolution performance

  • Record the spectrum of Polystyrene reference standard
  • Check the % absorbance at the absorption minimum at 2870 cm-1 and that at the absorption maximum at 2849.5cm-1.The difference (say X) in % absorbance between the two should be greater than 0.33
  • Check the % absorbance at the absorption minimum at 1589 cm-1 and that at the absorption maximum at 1583 cm-1.The difference (say Y) in % absorbance between the two should be greater than 0.08