Laboratory manager asking questions about GC to a female candidate during an interview
Image created in Bing

Interview questions on Gas Chromatography

Interview questions on Gas Chromatography (GC)

Gas Chromatography (GC) is one of the widely used tools to control the quality, safety and efficacy of pharmaceuticals at various stages. This is the reason why most of the questions are asked in pharmaceutical analytical or QC interviews. In this post I will discuss all the questions asked during the interview and their answers. If you want to clear the pharma interview then this post is for you.

GC is a chromatographic system to separate organic volatile molecules. In GC mobile phase is gas and the stationary phase is immobilized liquid. Separation is achieved by the interaction of molecules between the stationary phase (immobilized liquid) and mobile phase (carrier gas) and it depends upon both polarity (of both stationary phase and molecule) and the vaporization point of the molecules. Inert gas like Helium or Nitrogen is used as carrier gas.

You may like: Analytical Method Development and Validation in Pharma: Key Insights

  • The basic difference between GC and HPLC is that GC is used for testing volatile components whereas HPLC is used for testing non-volatile components. Some of the volatile components can be analysed in HPLC also.
  • In GC capillary columns are used whereas in HPLC packed columns are used.
  • In GC sharp peak is obtained compared to HPLC.
  • In GC higher theoretical plate (up to 100000 or more than that) compared to HPLC
  • In GC paek having less tailing factor  compared to HPLC
  • The basic difference between GC and TLC is that GC is used for testing volatile components whereas TLC is used for testing non-volatile components.
  • GC comes under continuous chromatography whereas TLC comes under complete chromatography.
  • GC is used for both qualitative and quantitative analysis of volatile components whereas TLC is used for mainly qualitative analysis of non-volatile components.

The different components of GC are:

  • Carrier gas
  • Flow control valve
  • Sample injection port
  • GC column
  • Column oven
  • Detector
  • Outlet
  • Computer controller

Most of the volatile components have no UV absorbance and hence they can not be analysed on HPLC using a UV detector. These can also not be analysed on other detectors due to volatility. However, these volatile components can easily be analysed on GC.

Answer: Carrier gas carries the volatile components.

 Generally, Helium and Nitrogen are used as carrier gas. These gases are inert which is the reason to use as carrier gas.

Answer: More than 99.5%

In GSC (gas-solid chromatography) packed columns are used whereas in GLC (gas-liquid chromatography) capillary columns are used.

GSC is not used due to the following reasons:

  • Broad peak
  • Low column efficiency
  • High tailing &
  • Poor separation

Due to fast analysis and high accuracy.

  • In the packed column low theoretical plate (N) whereas in the capillary column high theoretical plate (N)
  • In the packed column broad peak is obtained whereas in the capillary column-sharp peak is obtained.
  • In the packed column high tailing factor (T) whereas in the capillary column less tailing factor (T) &
  • In the packed column, poor resolution (R) is obtained between the peaks whereas in the capillary column higher resolution (R) is obtained between the peak

In the packed column, the reproducible result is not obtained and that is the reason the same is rarely used in the industries

Capillary columns are widely used in the industries compared to packed column due to the following reasons:

  • In the packed column low theoretical plate (N) whereas in the capillary column high theoretical plate (N)
  • In the packed column broad peak is obtained whereas in the capillary column-sharp peak is obtained.
  • In the packed column high tailing factor (T) whereas in the capillary column less tailing factor (T) &
  • In the packed column, poor resolution (R) is obtained between the peaks whereas in the capillary column higher resolution (R) is obtained between the peak

The sample is injected in the injection port. It is at a high temperature and it converts the sample into a vapour phase.

It helps in managing the column temperature as per method requirement.

Answer: Column is the heart of the GC system. In the column, different components are getting separated.

Answer: The detector converts eluted components into a signal and then the signal is converted into a peak by the data processor.

 Generally following detectors are used in GC:

  • Flame ionisation detector or FID
  • Thermal conductivity detector of TCD
  • Electron capture detector or ECD &
  • Mass detector or MsD

No

Yes

Because most of the organic compounds have oxidisable carbon atoms.

For halogenated carbon compounds like CCl4, CH3CH2Cl

The detector converts eluted components into a signal and then the signal is converted into a peak by the data processor.

Dimethylpolysiloxane, diphenyl dimethyl polysiloxane etc

The column is selected based on the polarity of the components.

Internal standard is used to nullify injection error.

Internal standard is used to nullify injection error.

For non-volatile molecule.

GC is used for both qualitative and quantitative analysis in:

  • Food industries
  • Pharmaceutical industries
  • Pesticide industries &
  • Petroleum industries

Non-volatile components can not be analysed by GC

First Ethanol will elute and then Isopropyl alcohol will elute.

  1. Ethanol
  2. Carbon tetrachloride
  3. Dichloromethane]
  4. Benzene

Answer: Ethanol and Benzene will be analysed on an FID detector. Carbon tetrachloride and Dichloromethane will be analysed on an ECD detector.

Answer: DB 624

Column bleed is the normal background signal generated by the column stationary phase.

Column bleed is the normal background signal generated by the column stationary phase

Column bleeding results from the thermal breakdown of the stationary phase at temperatures close to the upper-temperature limit of the column.

Factors affecting column bleed:

  • The level of bleeding depends on the film thickness of the stationary phase layer on the adsorbent. Thicker film columns are used to handle higher sample loads but also have a higher bleed rate. Bleeding at high temperatures is natural for all columns, regardless of the thickness of the stationary phase film, but excessive bleeding results in a shortened useful lifetime of the columns.
  • The presence of oxygen at higher column temperatures contributes to increased column bleeding and consequently increases the baseline. It is essential to ensure freedom from leaks in injectors and the use of oxygen traps in carrier gas lines.

Steps to minimize column bleed:

  • Always operate columns at least 20 – 30 degrees C below the upper specified temperature limit
  • Install oxygen and moisture traps in carrier gas lines and perform leak checks after column installation
  • Condition columns as per the manufacturer’s recommendations before use
  • Cap the open ends during storage
  • Replace septa frequently to prevent damage due to coring
  • Problems in both qualitative and quantitative analysis
  • The problem with column conditioning
  • Column life getting short

Column manufacturer manual must be followed while developing and finalising the GC method to avoid column bleeding. Use a low bleed column where analysis is required at higher temperatures. Low bleeding columns offer advantages such as the extension of the useful life of the column and improved
detection sensitivity at elevated temperatures. The advantages are of greater significance in high-sensitivity GC – MS analysis.

Gas Chromatography (GC) is a chromatographic system to separate organic volatile molecules. In GC mobile phase is gas (such as nitrogen and helium) and the stationary phase is immobilized liquid. Separation is achieved by the interaction of molecules between the stationary phase (immobilized liquid) and mobile phase (carrier gas) and it depends upon both polarity (of both stationary phase and molecule) and the vaporization point of the molecules.

Gas solid chromatography (GSC) and Gas liquid chromatography (GLC) are the the two types of GC

Nitrogen and Helium are used as carrier gas in GC

Carrier gas Nitrogen and Helium are used as a mobile phase in GC

Inert gases like helium and nitrogen that do not undergo chemical reactions under given conditions are called inert gases

Volatile compounds

Operating Principle of Flame Ionization Detector (FID)

  • The detector is highly sensitive to hydrocarbons and organic molecules.
  • FID detects ions produced during the combustion of organic compounds in a hydrogen flame.
  • When a sample elutes from the GC column and passes through the flame, carbon-based compounds (like hydrocarbons) are ionized, and the resulting ions produce a current that can be measured.

Operating Principle Thermal Conductivity Detector (TCD)

TCD measures changes in the thermal conductivity of the eluting gas relative to a reference gas.

As the sample components pass through a heated filament (the detector’s sensing element), any change in thermal conductivity (due to the different properties of the sample compared to the reference gas, usually helium or hydrogen) causes a change in the temperature of the filament, which is measured and used to generate a signal.

Difference

FeatureFID (Flame Ionization Detector)TCD (Thermal Conductivity Detector)
Detection principleIonization of organic compounds in a flameChange in thermal conductivity of sample compared to reference gas
SensitivityVery sensitive to hydrocarbons and organic compoundsLower sensitivity, detects a wider range of compounds
SpecificityHighly specific to organic compounds (especially hydrocarbons)Non-specific, responds to any compound with different thermal conductivity
ResponseLinear over a wide range, especially for hydrocarbonsLess sensitive, linear but for a broader range of compounds
ApplicationsOrganic compound analysis (e.g., hydrocarbons)General gas analysis (e.g., inorganic gases, non-organic compounds)
  • Liquid: It works with liquid samples
  • Headspace: It works with volatile organic samples, and
  • SPME: which uses fiber to extract the components of your sample

The length of the packed is 6 feet & length of capillary column is 10 to 100 meter in GC

In gas chromatography (GC), makeup gas is used to ensure proper detector function and to maintain a stable baseline. Its primary functions include:

  1. Maintain Detector Sensitivity: In certain detectors, like the flame ionization detector (FID) or thermal conductivity detector (TCD), the makeup gas serves to stabilize the signal by ensuring consistent flow and reducing baseline fluctuations. It helps maintain an optimal environment for the detector to function properly.
  2. Flow Rate Adjustment: Makeup gas can be used to adjust the flow rate to the detector, particularly when the flow from the column is not sufficient. This is common when using narrow-bore columns or when the detector requires a higher flow than what the column alone provides.
  3. Increase Signal-to-Noise Ratio: By improving the performance of the detector and stabilizing the overall system, makeup gas can enhance the signal-to-noise ratio, leading to more accurate and reliable measurements.

Makeup gas is typically an inert gas like nitrogen or helium, chosen based on the type of detector and the experimental requirements.

Zero air, as used in gas chromatography, refers to air that has been purified to remove specific contaminants, primarily any hydrocarbons or other reactive gases, that could interfere with the analysis of the sample. The composition of zero air typically includes:

  1. Nitrogen (N₂) – The primary component of air, nitrogen serves as an inert carrier gas in gas chromatography.
  2. Oxygen (O₂) – Present in the atmosphere and generally included in zero air, though in very low concentrations to avoid any reactive interactions.
  3. Argon (Ar) – A noble gas that may also be present in trace amounts.
  4. Carbon Dioxide (CO₂) – Often removed or maintained at a very low concentration, since it can interfere with the chromatographic process.

The air is typically treated through filtration and sometimes additional chemical scrubbers or catalytic converters to remove hydrocarbons, carbon monoxide, sulfur compounds, and other potential contaminants. The goal is to create an atmosphere that does not introduce interfering substances that could skew the results of the gas chromatography analysis.

In gas chromatography (GC), the “split ratio” refers to the proportion of the sample that is split between the column and the waste during the injection process. It is typically used when injecting a sample into the GC system with a split injector, which divides the sample flow between the column and the vent (waste).

The split ratio is defined as:

For example, if the split ratio is 10:1, it means that for every 1 part of the sample that goes into the column, 10 parts of the sample are directed to the waste. A higher split ratio reduces the amount of sample entering the column, which can help prevent overloading the column with too much analyte.

Split ratios are chosen based on the concentration of the sample, the sensitivity of the detector, and the capacity of the GC column

The following parameters are performed in GC calibration:

  • Column oven temperature
  • Flow rate of the gases
  • Detector performance
  • Detector precision and
  • Detector linearity

The main advantages of GC are fas analysis, cost effective and reliable result

More volatile molecule elutes first

The role of the GC (gas chromatography) oven is to heat the sample and the GC column to facilitate the separation of the sample components.

<621> Chromatography

Using reference standard you can identify peak in gas chromatography.

In Gas Chromatography (GC), spitless injection and split injection are two different modes of introducing a sample into the column, and they have distinct differences in how the sample is handled. Here’s an overview of each:

1. Split Injection:

  • Mode: In split injection, a small portion of the sample is injected into the GC, and the majority of the sample is diverted to the waste. This is done by using a split valve.
  • Split ratio: The amount of sample that enters the column compared to the amount of sample that is discarded is controlled by the split ratio. For example, a 1:10 split ratio means that 1 part of the sample goes into the column, and 10 parts are diverted to the waste.
  • Application: Split injection is typically used when the sample is concentrated or the analytes are present in high concentrations, and you don’t want to overload the column. It helps in minimizing the risk of overloading the detector.
  • Sample size: A relatively small amount of sample is injected into the system because most of it is vented out.

2. Splitless Injection:

  • Mode: In splitless injection, the entire sample is directed into the column without any diversion. This is typically done by closing the split valve during the injection so that all of the injected sample goes into the column.
  • Application: Splitless injection is useful when the sample is very dilute or when you’re analyzing trace levels of analytes that are difficult to detect. Since all the sample enters the column, it allows for better sensitivity and is ideal for low-concentration samples.
  • Sample size: In splitless injection, a larger volume of sample is injected into the system, and the sample is retained within the column for detection.

Key Differences:

  • Split ratio: Split injection has a split ratio (e.g., 1:10), while splitless injection uses a ratio of 1:1 (no splitting).
  • Sample handling: In split injection, most of the sample is discarded, whereas in splitless injection, all of the sample is transferred into the column.
  • Application: Split injection is preferred for concentrated samples, while splitless injection is ideal for trace analysis and low-concentration samples.

In summary, split injection is best suited for samples with higher concentrations, while splitless injection is designed for trace-level analysis where all of the sample is introduced into the column for better sensitivity.

In Gas Chromatography (GC), the liner plays an important role in the injector system. It serves as a critical component within the injection port, which is where the sample is introduced into the chromatograph. Here’s the specific role of the liner:

  1. Sample Introduction and Vaporization: The liner provides a chamber where the liquid or solid sample is injected. Once introduced, the sample is rapidly vaporized due to the high temperature of the injector. The liner ensures that the sample is evenly vaporized and then directed onto the column.
  2. Improved Sample Mixing: The liner helps mix the sample with the carrier gas, aiding in efficient vaporization and helping to prevent sample degradation or degradation of thermally sensitive compounds.
  3. Preventing Contamination: Liners help prevent the sample from directly contacting the injector’s metal parts, which could otherwise cause contamination or adsorption. It acts as a barrier, ensuring that the sample does not react with components of the injector.
  4. Reducing Peak Broadening: The liner is designed to promote a smooth flow of vaporized sample into the column, preventing the introduction of impurities that could cause peak distortion or broadening in the chromatogram.
  5. Maintaining Reproducibility: Since the liner is a part that comes into contact with each sample, its material and cleanliness can influence the precision and reproducibility of results. Regular cleaning or replacement of the liner is necessary for consistent results.

In short, the liner ensures efficient sample vaporization, minimizes contamination, and helps maintain the integrity of the chromatographic analysis.

Then separation will be decided by polarity of the sample and the column.

The time at which analte peak elutes in GC is called retention time

Temperature plays a very important role in peak separation and retention time optimization.

In injection port sample is injected and it is at higher temperature.

The purpose of using guard columns is to protect the GC column from contamination since the sample that is introduced may be impure like reaction mixture.

Resolution in Gas Chromatography can be optimised by oven temperature programming, carrier gas flow rate by split ratio.

Carrier gas must be inert to avoid any reaction with sample components.

FID (Flame Ionization Detector) is widely used in Gas Chromatography since most of the organic compounds contain oxidisable carbon atom.

Using mass detector in GC molecular mass of unknown analyte can be determined and structure can be characterised.

Once in six months.

Conclusion

I hope this article has cleared all douts related to GC interview questions and now you answer all GC related question in the interview.

You may also want to check out other articles on my blog, such as: