Relative Response Factor (RRF) and its Calculation in HPLC
Table of Contents
- Introduction and Background
- Method for calculating impurities
- Result of impurities using different methods
- Method for calculating the RRF
- Relative response factor calculation by Response ratio method
- RRF calculation by slope method
- Factors affecting RRF
- RRF of Isomeric impurities
- Can RRF of unspecified impurities be calculated ?
- Conditions for RRF Considerations
- Advantages of RRF
- Conclusion
- FAQs
Relative Response Factor (RRF) plays a unique role in pharmaceutical analysis. The quality of Active Pharmaceutical Ingredients (APIs) or their stages depends on their impurity profile or in other words, we can say purity. Impurity standards are required to test the impurity profile of APIs or pharmaceutical products or their stages. However, given the cost, analysis time, and availability of impurities standards, it is impossible for any pharmaceutical company to manage each impurity and use each impurity during routine analysis. Now the question is what is the solution? The only solution is the use of Relative Response Factor (RRF) and that is why I decided to share my skill-based knowledge on “Relative Response Factor (RRF) and its calculation in HPLC and GC analysis“.
In this post, you will learn about RRF (Relative response factor), Response factor, need for response factor, method of calculating impurities, case studies, procedure of calculating response factor and FAQs. This article will clarify all your doubts related to RRF

To deepen your understanding of Relative Response Factor, be sure to check out this article: Relative Response Factor (RRF) in Pharmaceutical Analysis
Relative Response Factor or RRF
Relative Response Factor or RRF is defined as the ratio between the Response Factor of impurity and the Response Factor of the main analyte standard :
RRF= Response factor of Impurity/Response factor of the Main analyte
Method for calculating impurities
The following techniques are used to calculate the impurities in the impurity profile or related substances test:
- Area normalisation method or area % method
- Against a diluted sample of the main analyte
- Against corresponding impurity standard or external standard method
Area normalisation method or area % method
The following formula is used to calculate the value of impurity:
% Impurity = (Area of impurity in the sample chromatogram÷Sum of area of all peaks in the chromatogram)x100
Against diluted sample of the main analyte:
In this method diluted Analyte standard is used to calculate the impurities
Against corresponding impurity standard/external standard method:
In this method, each impurity is calculated against its corresponding impurity standard.
Result of impurities using different methods
In figure-1, sample analyte A contain two impurities B and C. Now let us calculate each impurity B and C using all three techniques.
Figure- 1 (Value of impurities calculated using different techniques

Result of impurities B and C calculated using different methods
Peak | Area | % value using Area normalisation/% area | % value using diluted Analyte | % value using external standard |
A (analyte) | 9878863 | 97.98 | – | – |
B (Imp) | 155245 | 1.54 | 1.49 | 0.62 |
C (Imp) | 48695 | 0.48 | 0.44 | 0.41 |
Total | 100280 | – | – | – |
Table-1
Challenges with the different methods for calculating the impurities
It is clear from the above (figure -1 and table-1) that the results obtained by the area normalization method and the diluted standard methods are almost identical for impurity “C” while the results obtained by the external standard method are different for impurity B.
Now the question is which method is correct out of area normalization, dilute analyte method and external standard method. In other words, which of the above results is correct?
The answer is result obtained by external standard is correct using all detectors. The result obtained by the Area normalisation method and diluted analyte method only be correct with the universal detectors like ELSD, LC-MS. GC-Ms and not by the UV detector.
For UV detector:
- The result depends upon concentration as well as absorbance of the molecules
- The absorbance of a molecule depends upon its structure as well as the wavelength used
As each impurity may have different wavelength maxima and hence, the result obtained by the Area normalisation method and the diluted analyte method is not correct.
The result obtained by the external standard method is correct.
But now the question is, can an external standard method be possible to use every time?
The answer is no because of the following reasons:
- The external standard method needs each individual impurity standard while performing the analysis.
- Most impurity standards are not commercially available.
- Impurities preparation/isolation and characterization is costly as well as time-consuming.
- It increases the cost of the project many folds.
- If the impurity is highly hygroscopic or unstable in that case, it is impossible to preserve and use as an external standard.
Now the question is what is the solution?
To get rid of the above challenges related to calculation by different methods, the Relative Response Factor (RRF) has been introduced. It is defined as the ratio between the Response Factor of impurity and the Response Factor of the main analyte standard :
RRF= Response factor of Impurity/Response factor of the Main analyte
And the following formula is used for the calculation of impurities:
% Impurity = (Area of impurity in the sample chromatogram/Sum of area of all peaks in the chromatogram)x100
RF approach is very helpful and acceptable by all regulatory agencies:
- USP uses F = RRF. It is used in the denominator for the calculation
- PHEur uses CF, where CF = 1/RRF, It is used in the numerator for calculation
- Generally, RRF is used in the denominator for the calculation
RRF Calculation Method
The following methods are widely used in the Pharmaceutical industries for calculating Relative response factor (RRF):
- Response ratio method
- RRF calculation by slope method
Relative response factor calculation by Response ratio method
This is a tentative method for calculating RRF and is generally used during the development stage. Impurity and Analyte standards are prepared at the same concentration and injected into a chromatographic system to get the chromatograms. RRF is calculated by the following formula:
RRF= Response factor of Impurity/Response factor of the Main analyte
RRF calculation by slope method
The relative response factor will be determined by dividing the slope of specified impurities by the slope of the reference substance

Procedure for calculating RRF:
- Prepare at least 5 concentrations from two different stock solutions for each impurity as well as for the main analyte at the concentration range QL (quantitation limit) to ≥ 150%
- Generate the chromatograms for each concentration for each impurity as well as for the main analyte using the test method
- Calculate the slope using using Excel for each impurity and API/main analyte.
- Calculate the RRF
RRF calculation case study
Let us consider an API which contains impurity “A” with a limit of NMT 0.075%. QL of the method is 0.0075%. The sample analysis concentration is 3000mcg/ml (3mg/ml).
Procedure
- Prepare at least five solutions from two different stock solutions for each impurity “A” as well as for the API between QL to 150%
- Generate the chromatograms for each concentration for each impurity as well as for the main analyte using the test method
- Calculate the slope using Excel for each impurity and API/main analyte.
- Calculate the RRF
Impurity A and API solutions preparation
- The sample concentration is 3000 mcg and hence impurity concentration at the specification level will be 3000 x 0.075/100 = 2.25mcg/ml.
- It means 2.25mcg/ml solution is equal to 0.075%. Now other solutions can be prepared in the following ways:
- Impurity A concentration QL (10%) will be 2.25 x 10/100 =0.169 mcg/ml
- Impurity A concentration 25% will be 2.25 x 25/100 =0.56 mcg/ml
- Impurity A concentration 50% will be 2.25 x 50/100 =1.125mcg/ml
- Impurity A concentration 75% will be 2.25 x 75/100 =1.69 mcg/ml
- Impurity A concentration 125% will be 2.25 x 125/100 =2.81 mcg/ml
- Impurity A concentration 150% will be 2.25 x 150/100 =3,38 mcg/ml
Note: Similarly API solution can also be prepared
% Concentration | API Concentration (mcg/ml) | API AREA | A Concentration (mcg/ml) | A AREA |
QL | 0.17 | 5030 | 0.17 | 31555 |
25 | 0.56 | 12360 | 0.56 | 8275 |
50 | 1.125 | 22960 | 1.125 | 16849 |
75 | 1.69 | 43941 | 1.69 | 33055 |
125 | 2.81 | 55422 | 2.81 | 42070 |
150 | 3.38 | 66390 | 3.38 | 50140 |
Slope | 19224.64333 | NA | NA | 9675.245121 |
RRF of A | (9675.245121/19224.64333) = 0.50 | NA | NA | NA |
RRF of A = Slope of A /Slope of B = (9675.245121/19224.64333) = 0.50
RRF for A is 0.54 and this will be used while calculating this impurity in routine analysis.
Factors affecting RRF
Relative response factor or RRF may change due to a change of:
- Wavelength)
- Column brand
- Column stationary phase particle Size
- Detector ( UV and PDA)
- Solvent grade
- Buffer concentration
- Column temperature
- pH variation
- Structure of the molecule
RRF of Isomeric impurities
RRF of Isomeric impurities may or may not be same.
Case study : ortho Benzaldehyde, meta hydroxy Benzaldehyde and para hydroxy Benzaldehyde are not identical RRF or RRF equal to 1
Can RRF of unspecified impurities be calculated ?
If unspecified impurity is known then RRF must be calculated and considered in impurity calculation. IF unspecified impurity is unknown then RRF can not be calculated.
Conditions for RRF Considerations
RRF between 0.8 and 1.2 should be considered as 1.0 and may not be used for calculation
- The RRF below 0.8 should be expressed with two significant numbers
- The RRF of less than 0.2 or more than 5 should not be used calculation
Advantages of RRF
- Helping in giving fast and exact results of impurities in drug substances, drug intermediates and drug product
- Universally acceptable e.g., by guidelines, by Pharmacopeia
- One-time evaluation job and lifetime calmness
- Avoid the stability/storage/management issues with standards
- Reduce the analysis/project development cost drastically
- Accepted by all Regulatory agencies
Conclusion
Relative response factor or RRF play a vital role in calculating the impurities by chromatographic techniques. and hence clear understanding is required for effective method development. This is all about this article. I hope now all your doubts related to RRF have been cleared and you will be able to use this technique in chromatographic analytical method development and routine analysis. Any questions related to this article write in the comment section and I will answer on a priority basis.
You may also want to check out other articles on my blog, such as:
- Difference between Development and Validation
- How to develop a method using GC-MS?
- Need of Chromatographic Method in Drug Development
- Allowable GC Method Adjustment
FAQs
What is the relative response factor?
Relative response factor (RRF) is used to control the impurities by chromatographic methods like HPLC and GC. In this method impurity standard is not used during the analysis and in place of impurity a RRF is used in the calculation.
How to calculate response factor?
There are several methods for calculating the response factor (RRF) in which the slope method is widely used in the industries. To calculate the RRF by this method, slope of impurity is divided by the slope of the main analyte.
RRF = Slope of an impurity/Slope of the API
Abbreviations:
- HPLC: High performance liquid chromatography
- GC: Gas chromatography
- RF: Response factor
- RRF: Relative response factor
- MS: Mass spectrophotometer
- ELSD: Electrospray detector
- QL: Quantification limit
References
- USP
- EP
- WHO guideline
Note:
- Some of the data have been designed to explain the topic
Very informative blog thanks
Keep reading and sharing